those probes do respond to light-especially quartz halogen- (could be the IP of the UHP gas is around the energy of quartz halogen energy) First change out the probe. if your problem clears then its the probe- have seen lots of problems with light on those detectors-we had a discussion about it a while back- when you have the problem isolated to the probe you have several options- gas purity problem, moisture build up in the probe, mylar not opaque, discharge corona point, carbonization of the anode,-(someone touched it and left body oils on it ) cold soldier joint, debris in detector, wrong type coax , damaged connector. those probes are pretty simple, also they dont do well in cold. thats about as instrument specific as anyone can get. the voltage plateau should be checked for that specific instrument package.