Well assuming you are "on contact" with the RO-2 (~1 cm since you dont want to touch the particle to the detector), and assuming its Co-60 (probably overestimation of energy), my guess would be if you still had 1mCi, it would overrange the RO-2 at just shy of 13 REM/hr....just my SWAG though since you didn't provide specifics like what nuclides (even if you provide a plant it came from doesnt mean we know what's in it), when the activity was measured (to decay correct), etc.....
If your'e looking more for like a plant mix, I would assume closer to 1 MeV gamma energy (I don't work at a PWR though so I'm not that knowledgable on what they have as an average mix), and roughly divide the dose by 2.5 (total gamma energy for Co-60 rounding yield for both gammas to 100%), which would give around 5 REM/hr.
Oh yeah, that's closed window.
I'd say your instrument would be good if you were within 20% of scale ;-).
Did that help?