NRC issued a generic letter way back in the early '80s on the requirements for free-release monitoring using hand held detectors. I would have sworn that it was Generic Letter 83-06, but looking through the listing at the NRC web site, I don't see it.
Do any of you remember that generic letter? If someone finds it, please let us know.
I know it existed because I used to keep a copy of it on file. I used it as a reference to explain to my supervisors that, even though our sophisticated contamination monitors "can" see activity at much lower levels, we don't want to do that because to do so would exceed the standard set by the NRC in the generic letter. We would be shooting ourselves in the foot to look for radioactivity at levels lower than the NRC's standard. That's why we dont use a GeLi to release everything.
The generic letter specified that the HP-210 detector was the standard for monitoring for free-release (<100 cpm), when held at a distance of no more than 1/2 inch from the article being monitored, and when moved no faster than 2 inches per second. The generic letter was based on a study done by some college prof. who published his findings in some journal, which was later cited by the NRC in the generic letter as the basis for setting the standard of "how hard you have to look".
Anyway, either the generic letter or the college prof's paper may also discuss contamination monitoring for pure beta emitters like Sr-90. Maybe that's where your 1.5 inches per second comes from.