RDTroja:

--- Quote from: makua13 on Feb 23, 2005, 03:32 ---I been calibratin them things for years and never saw 18%, not with Tc99 anyways.  Funny how numbers get larger when you get older..  ::)

--- End quote ---

Interesting... just the other day I was thinking that some numbers get smaller as you get older...

And I never referenced Tc-99. Not sure I ever used that as a calibration source myself, but that was 100 years ago or so.

RDTroja:

--- Quote from: 870xprs on Feb 23, 2005, 07:09 ---... even though our sophisticated contamination monitors "can" see activity at much lower levels, we don't want to do that because to do so would exceed the standard set by the NRC in the generic letter.  We would be shooting ourselves in the foot to look for radioactivity at levels lower than the NRC's standard.  That's why we dont use a GeLi to release everything.
--- End quote ---

I wonder what happened to that concept. If I ever get my hands on the idiot that came up with 'no detectable activity' as a limit I think I may strangle him.

And the first time someone in the position of authority recommends using a GeLi to free release something, I am coming after you, 870xprs, just for putting it in print!  ;)

Shonkatoys:
The  limit  for  Strontim  is  only 1000 dpm  100 cm2 where  the  limit for  uranium is 5000 dpm  100 cm 2  therefore  I  would  slow  down even more than 1.5  inches  a   second  to  realistically  check for strontium per 10cfr 835  to  detect  the  lower  limit.
  For  transuranics  I  was  told  you  should  frisk at 1  inch  every 11 seconds  with  a  43-5 ZnS probe to  meet  the  limits,  However  no  one  does  this.  Transuranics  should  only  be  surveyed  with  a   large  surface  Contamination  monitor  if  surveying  buildings in  my  opinion.  Surveying  for  transuranics  is  a  joke  on  everysite  I  have  been in  my  opinion.  20 clicks  different  from  background  give  or  take a  few  clicks in  a  minute.  You  must  be  crazy  if  you  think  that  anyone  can  survey  to  this.   Too  easy  to miss 500 dpm  100 cm 2. 

stownsend:
There are a couple of limits for Sr90,either pure or in a mixed fission product.When you take into consideration of the energy,eff,and probe size your only looking for ~30 counts so you need to stress the turbo frisk doesn't work or as an HP help frisk them.As far as transuranics go you can't see  20 dpm/100cm2 in the field,only on lab equip.ie.tennelec or protean and frisking to the 500dpm/100cm2 total takes a trained eye .That's why we seniors make the big bucks!!!

870xprs:

--- Quote from: RDTroja on Feb 23, 2005, 09:34 ---I wonder what happened to that concept. If I ever get my hands on the idiot that came up with 'no detectable activity' as a limit I think I may strangle him.

And the first time someone in the position of authority recommends using a GeLi to free release something, I am coming after you, 870xprs, just for putting it in print!  ;)

--- End quote ---

That's ok, but I should point out that it's considered one of the rites of passage for Sr. HPs that they have to 'splain things to someone in a position of authority.  That's another reason they pay us the big bucks.  ;)   ::)