Reference, Questions and Help > Instrument Q&A
What calibration sources do you use for dose rate instuments?
radgal:
What are the sources used out there to calibrate Beta/Gamma dose rate instuments?
Also what do you use for daily reponse checks, source and strength?
There are just to satisfy my curiosity what different places use.
Thanks
jjordan:
We use the Shepard to calibrate with, and we also source check the Tele's and AMP 100-200's with it. For the RO's we use a Sr 90 2.0 e +4 uCi source with differnt shields to source check on all scales. I don't know if you got to see it or used it when you were here at Brunswick.
JJ
makua13:
C137 and Tc99 are the standards used by most (power) plants, but depending on what you are looking for the source may change.
SloGlo:
--- Quote from: Rad Gal on Jun 22, 2005, 11:39 ---What are the sources used out there to calibrate Beta/Gamma dose rate instuments?
Also what do you use for daily reponse checks, source and strength?
There are just to satisfy my curiosity what different places use.
Thanks
--- End quote ---
i've used cs137, tc99 most often, usually due to it's long life yielding less frequent t1/2 calcs. on rare sites, i've used co60.
most are small content, 0.05 mci or so. some have been bigger. most are ~1" dia, but there have been rare times when bigger (~2") where on the site.
radgal:
SloGlo, Makua13 and JJordon
Thanks for the replies. This is what I thought you guys used. I just looking for confirmation
JJordon: Do you have a Beta correction factor for your RO20 due to your calabration source being a really strong Beta emitter. (According to Eberline the RO20 is 93% efficienct for SR90 but only 30% efficient for Uranium Slab detection.) I believe I know the answer but since I worked there before but I would just like it for the record, so to speak.