First off,
I have a 150 mCi Cs-137 Shepard calibration source I'm working with and is normally used to calibrate GM's. However, I've been asked to attempt calibration of our Keithley Ion Chamber model 36150. I want to achieve 1.0 +/- 20% on all three scales.
Can anyone give me some calculations or ideas on the best method to calibrate this ion chamber? This is a basic calbration and not to be held to ANSI standards.
Thanks in advance
-
Well, it seems I have no replies. Does anyone have a copy of the manual for the 31650 they could scan and send me? If you have one send it to cvande2@emory.edu
Anyone know who owns Keithley?
-
Perhaps no one understands your question.
There is no difference in calibrating a gm or ion chamber to gamma in a shepherd.
Place the detector in the proper geometry and distance at what ever value you know and adjust the meter to read what it should at that distance.
For example I have used a shepherd that I know the table dial setting of 200 give me a dose rate of 250 mr/hr I set the instrument there and adjust as needed. This calibrates the instrument to gamma only. From my experience to use it for beta also you need a beta source with a known strength at a known distance. You take a reading and then determine a beta efficiency and appropriate multiplication factors.
open window-closed window times beta correction factor give you a beta dose rate.
I have never heard of the instrument you mention but I have seen the process I describe used at various nuclear plants.
Hope that helps
sf